I have a device which can be modeled with a resistor in parallel with a capacitor. Now I want to measure its resistance and capacitance as a function of voltage bias (from -2V to 2V) and frequency (from DC to 10kHz). Originally I was thinking of using a LCR meter such as hand held Agilent U1732C which can measure in 100Hz, 120Hz, 1kHz and 10kHz. But the problem is I am not sure how to apply a bias and measure with this equipment at the same time. On the other hand, I don't need very high precision. I think 5% precision should be okay. But more frequency points from DC to 10kHz will be helpful. So my question is:

  1. How can I apply bias on the device and measure its C and R using Agilent U1732C? Can I do that?

  2. If 1 doesn't work, can I do it with a not-so-complicated DIY kit rather than resorting to a high-end LCR meter?

Best Answer

If it were me , I would not only measure C & tan delta but ESR as well and test to 10MHz.

You can use a scope rear sweep signal to drive an FM generator, but 50 Ohm is not the best source.

If you cant scrape a simple AC current source design to generate cap impedance as the output voltage, use a voltage source (ie. darlington emitter follower to drive the cap with an adjustment for DC offset +/-2 V and monitor current with a 1 Ohm shunt. The scope can show the DC bias on one channel and ac coupled envelope synced to scope sweep. This is a quick & dirty method. XY mode of V vs I is another method.