Electronic – Current measurement on low resistance sources

current measurement

Most of the devices to measure current at micro amps level or lower recommend a input resistance of more than 1kOhms.
(As for an example , shorting the ammeter leads, which is essentially what happens on connecting the low resistance source, produces noise in 0.5 mA range)

How do we measure current from sources which have very low resistances?

Best Answer

One way to measure very low resistances is to pass a relatively large current through the device under test (DUT), with a Kelvin (4-wire) connection. Then a precision amplifier can be used to measure the voltage drop.

http://img.deusm.com/designnews/2013/10/269248/1-fig-1.jpg

For example, suppose you have a 1A current source and an amplifier with 10mV full scale input (and say a 3-1/2 digit ADC). Then you can measure with 5\$\mu\Omega\$ resolution.

It may be better to use +/-1A alternating and make alternate +/- measurements in order to get rid of thermoelectric voltages, but that's a detail.