Disclaimer: I'm not a EE professional (I have a background in physics though), so please bear with me if the question seems banal, obvious or just plain stupid.

We have a Keysight 34465A DMM, which, according to the data-sheet, has DC-I measurement accuracy of 50 pA (typ) at it's lowest range of 1 uA when the input is close to zero.
When the device is set to this setting (auto-zero enabled, 10 PLC) and nothing connected, the display fluctuates within a few pA around zero. As soon as a probe cable is connected to the common port (labelled "Input LO"), the display shows a measurement value of between +2.5 and +3.0 nA DC. If it is connected to the current sense port (labelled "Input 3A"), it shows values of around -50 to -150 nA DC! The probe is a simple passive test lead (Keysight 34138A). If two such leads are connected to these two ports, the displayed value remains in a similar range, unless they are shorted, where the measured current is again within a few pA around zero.

To me, i's absolutely puzzling to see any DC current beyond the specified accuracy of the device when "it is not connected to anything". The best I could come up with is an AC current due to EMI that somehow reduces the device's accuracy via the induced noise. However, the values seem very high to me.

So here are my questions: What is the origin of these currents? Is this to be expected or is it a hint to a problem with the device or the workplace? In view of those issues, how do I measure currents with nA accuracy coming from a high-impedance source?

Best Answer

What is the origin of these currents?: Every cable is an antenna. I think you measure noise (FM radio, cell phone, WIFI, Bluetooth, DECT, ...)

How do I measure currents with nA accuracy? Use shielded cables; shield your DUT; Use a battery as supply for your DUT; Eventually use a shielding enclosure for the whole setup (DUT + DMM); increase the number of samples the DMM uses to build the RMS value;