How accurate can I get temperature measurements if I repeatedly sample the data from a PT100 sensor. [Assuming I have a 16 bit converter]?
What are the limits on accuracy and repeatability? Does the accuracy increase as the square root of the number of samples?

Best Answer

Limits include the sensor accuracy (calibration and interchangeability at the temperatures of interest), self-heating of the sensor, coupling of the sensor to the thing you're trying to measure, effect of the sensor on the thing you're trying to measure (heat input or heat loss, for example), instrument calibration and changes in that with time and environmental factors, and noise.

Noise (including quantization noise) is the only one that you can improve by averaging measurements, and not necessarily even that, depending on the type of noise.